Dear @dbozzato
Please find attached the reference article.
portable Am-Be source.pdf (735.7 KB)
I have been using the user routine package given in Sampling from energy spectra without writing my own.
Thank you
Dear @dbozzato
Thank you
Dear Vincent,
thanks for sharing, now I have a clearer picture.
If you want to make a comparison, the first step would be to use the same spectrum as in the article you shared (or a reasonable approximation of it) for the sampling of the neutron primary energy. If I compare the one in the article to the one in the file you shared, there are substantial differences in the spectra features and in the granularity: the absolute values don’t matter for the sampling but of course the shape and presence/absence of peaks does.
As far as the dose conversion coefficients is concerned, the ones in the article. once converted to pSv cm^2, are at first glance not significantly different from those currently available in FLUKA so they shouldn’t be the source of your issue: with WHAT(2) of the USRBIN equal to DOSE-EQ you score ambient dose equivalent (H*(10)) by default but you can use the AUXSCORE card to change the sets and score effective dose as well. The sets EAPMP (effective dose for anterior posterior irradiation) or EWTMP (effective dose for worst possible irradiation condition) are both valid options (the latter more being more conservative)
If you will need to further refine your comparison, remember that FLUKA also allows to apply user defined coefficients using the fluscw routine.
Davide
dear @dbozzato
Thanks for the response. I must mention that the reason for the difference in spectra shape is that the one I used is expressed as iso-lethargic fluence from the IAEA spectra compendium, so it should be different from that in the article for which I don’t have data. Therefore, I will have to use the article that employed accessible spectrum data.
Thank you @sunil for sharing.
@Motlatsi_Vincent, if you encounter further problems, do not hesitate to ask!
Davide
The ISO spectrum was helpful. However, I still think I need the one whose shape matches the spectrum in the reference article.
Thanks so much
You can extract the plot data from the publication using a digitizer. One such app is
https://apps.automeris.io/wpd/
I have extracted values from the reference article using the Web Plot Digitizer. I checked the similarity of the spectrum shapes by producing the spectrum in FLUKA using Y to get the spectrum in terms of dN/dE. I see different shapes and relative intensities. I also get large dose rates that are above 100 uSv/h when the target dose rate is any value below 25 uSv/h.
How do I make sure I reproduce the same spectrum to compare the results of dose rates? Please suggest
here I provide the extracted data
extr.dat (2.0 KB)
Dear @Motlatsi_Vincent,
it is not very clear to me how you reproduced the spectrum in FLUKA. Could you please share the flair project and all the relevant files that you have used in your simulation? It would be much easier to advise then.
Many thanks,
Davide
Dear @dbozzato
Here I provide the Flair file
Spec.flair (2.1 KB)
Please note that I did not use source.f routine but I followed instructions in Sampling from energy spectra so that I could use the spectrum data.
After extracting data, I rearranged the energy values into energy bins as used in the ASCII file. The values in the last column are the intensities corresponding to the intensity values on the spectrum in the reference article. They were fed to the ASCII file without changes.
In FLUKA, I used the USRBDX card to score the neutron spectrum in linear energy of neutrons escaping the spherical region surrounding the source. In Flair, I chose Y to get a linear spectrum instead of a logarithmic one.
Hi @Motlatsi_Vincent
I am not sure if you got a reply to this, but I am able to reproduce the original spectrum exactly, using the data file you provided and the source.f that you pointed to. Change the energy binning to match the input spectrum range to get a clear plot ( I used 1e-4 to 0.1 GeV ), and, replace iron and air with vacuum (you are testing your source, so don’t modify it by introducing any material). While I used fluence (WHAT(1)=111) ( you had it as current) it may not matter with these changes.
You should get this. Spec_plot06.pdf (19.1 KB)
Sunil
Dear @sunil
Thank you for the response.
I managed to reproduce the original spectrum in FLUKA through your suggestion, as you did too. Our spectra, indeed, have similar shape and peaks to the one in reference article, but I thought the emission probability (in MeV-1) would be the same. Anyway, using this spectrum, I estimate the dose rate outside the tank designed as stated in one of my questions above, using the flair file in my first question. I still get high dose rates. Can the different emission probability be the reason for high dose rates?
Your suggestion will highly be appreciated.
I gave a cursory look at your original FLAIR file and the article that you had attached. Perhaps these points may help you cross check and improve (Caveat: I did not read the full article.)
Cross check the normalization of FLUKA results. FLUKA outputs are in pSv/primary. Use the same source strength (n/s) as given in the article and make sure your normalization gives you the same unit as in the article (mSv/h). I suggest you double check this to start with.
Check to see if the source was implemented as a volume source or a point source in the article. If it is a volume source, you will have to implement such a sampling also in the source.f for FLUKA.
Check to see if you got the geometry/materials correct. Check to see if you have the composition of the materials correct. I think there are some variations, but again, I might be wrong since I did not read the entire article.
There is also the fluence to dose conversion coefficient factor, but @dbozzato has already cross checked that, so you can rule out that a reason for the discrepancy for the time being.
Good luck!
apologies for the late reply.
As @sunil pointed out, the spectrum sampling is correct and from a first glance the main source of discrepancy is in your geometry. You have implemented only an iron tank but, with respect to the article, you are missing the tungsten layer after the source (which is approximately cylindrical and not point like) and the borated polyethylene that acts as absorbing material. These two elements cannot be neglected at all!
You can try to reproduce as much as possible the geometry as in figure 5 of the article or using the geometry information the authors provided for M and L. This should guide into the right direction.
Let me know if you encounter issues in the next iteration.
Davide
The iron tank was for testing the spectrum reproducibility.
I defined the geometry that is similar to their geometry and assigned the materials to regions accordingly ( materials: tungsten and borated polyethylene)
To ensure dose rate unit correspondence, I preferred to work in uSv/h and used their neutron emission rate and the conversion factor as 3600e-6 so that the dose rate limit is 0.025 mSv/h = 25 uSv/h.
I still get the high dose rates. I doubt that the borated polyethylene is correctly defined. They stated that the absorber is polyethylene plus 1.2% boron carbide (PE- 1.2 % B4C).
Please assist in defining this material, density, in particular, as I doubt it is correct, and it possibly impacts the calculations. Maybe it can solve this issue.
Here I provide the flair project.
Code-Validation.flair (3.2 KB)
extr.dat (2.0 KB)
Dear FLUKA users
I am reproducing the neutron shielding results obtained with the MCNP code in one article (check the article in Comparison of results for validation should there be a need).
So far I have defined a similar geometry, and I use a similar energy spectrum for the 241Am-Be neutron source.
The problem is with defining the material composition of Polyethylene containing 1.2 % boron carbide (I doubt I got the density of the composition correct).
The ultimate aim is to get the dose rates less than 0.025 mSv/h (=25 uSv/h) outside the geometry if everything is well defined.
Can you please assist in defining this material? Here I provide the flair project and the spectrum data. I used routine provided in Sampling from energy spectra
extr.dat (2.0 KB)
MCNP-results.flair (3.3 KB)
I gave again a careful read to the article and tried to follow the references. Apologies for what will be a long answer. Let’s tackle the main points:
Please have a look at the flair file and give it a go. Please let me know in case you have any question
MCNP-results_suggestions.flair (8.0 KB)
All the best,
Davide
For info, I re-checked the dose coefficients: originally I had given a detailed look to the neutron ones because they were from an American Standard which was unknown to me and as I replied to you the differences were quite small although the conversion coefficients implemented in FLUKA have a much more refined granularity. I trusted that the photon ones were close to the ones in FLUKA since they were from an ICRP publication.
Well, I went back to the quoted ICRP publication, in particular to table 3 at page 16: the coefficients in the ICRP publication were given as (photons/cm2 s)/(mrem/h) so then ones in the article should be their inverse divided by 100 for the conversion rem->Sv. The coefficients in the article, though, are a factor 10 less than what should be!
In this case this has a small effect since the dose due to photons is much smaller than the one due to neutrons at least laterally to the tank. I have tried using the coefficients in the article in a routine and the difference with respect to the FLUKA built-in scoring is around 20% laterally outside the tank.
Please let me know if you have further questions.
Davide