Patient QA DOSE Verification

Dear FLUKA Users and Experts
Hello everyone, Can Fluka bu used for dose verification of patient QA(Particles:Photon), If it can be used for Photon dose verification,Is there any relevant operation explanation.
I am not familiar with the relevant operation settings.

Cheers,
Chailin

Dear @CHAILIN

yes you can use fluka for photon QA. flair v3 is capable in converting photon plans into inputs that fluka v4 can understand and run. The only problem is that the normalization of these plans are reported in MU (monitor units) and not in NP (number of particles) therefore you need to create a custom beam model in the RTPlan of flair that converts the MU to NP. This beam model is different for each facility/setup.

Cheers
Vasilis