Dear FLUKA team,
I am performing a 6 MV photon 3D-CRT treatment plan recalculation for a real brain patient using FLUKA.
I performed a simulation of the clinical LINAC and obtained a dose of:
x=9.3×10⁻⁷ Gy/primary.
According to the procedure I am following, I calculated:
NP/MU = 1/(100*x)
which gives approximately:
NP/MU=10753 primary/MU.
I then used this NP/MU value in the Beam Model File for the patient treatment-plan simulation.
I would like to confirm the following:
-
If x is already expressed in Gy/primary, is the calculation
NP/MU = 1/(100*x)correct? -
Since x is already in Gy/primary, is it correct that no additional conversion factor such as 1.602×10⁻⁷ is required?
-
Where should I obtain the number of irradiation particles from? Is it obtained from the Beam Model File, the clinical MU, the RT Plan, or somewhere else in the FLUKA simulation?
-
What is the recommended equation for converting the FLUKA dose per primary into the absolute patient dose in Gy?
For example, if a patient field has MUi, should the number of primary particles be calculated as:
N_i = (NP/MU) × MU_i
and then calculate the absolute dose as:
D_absolute = D_FLUKA/primary × N_i
Or is this conversion already handled automatically by the Beam Model File, in which case applying this multiplication again would result in double normalization?
I would also like to clarify this because, in my patient simulation, I obtain a CTV dose from USRBIN DOSE of approximately:
4.0341×10⁻⁷
while the corresponding TPS dose is approximately:
31.592 Gy.
I want to make sure that I am not applying an incorrect additional normalization factor when comparing the FLUKA DVH with the TPS.
Thanks in advance very much for your help and clarification.
Best regards,